Thanks for the replies! I'll bring up the possibility of testing the testers to the bosses and see if they're open to it.
Thanks for the replies! I'll bring up the possibility of testing the testers to the bosses and see if they're open to it.
OK, I reached the person who had the info, and unfortunately, there was not one definitive report.
1. The most detailed study (1,000 pool water samples with 6+ kit types) was done informally by Que Hales of PoolChlor several years ago. I should have the news report soon, but it may be 6 months (after the pool season!) before I the actual test data.
2. Another study, paid for by the pool plasterer's trade association, was done at the Cal-Poly pool center, but it costs $500+ to get a copy, and does NOT include data or methods, only conclusions.
All chlorine indicators (OTO, DPD, syringaldazine) have some interferences. The use of a non-chlorine shock (often potassium monopersulfate) will always produce a false combined chlorine reading, and will sometimes, I seem to recall, produce an erroneous free chlorine reading.
Richard Falk (Chem_Geek, here and on other forums) probably has detailed info on this -- I'll ask him to look at this thread.
PoolDoc / Ben
The titration method is only accurate to within one drop which for a 25 ml sample size would be 0.2 ppm. When you add add the last drop that extinguishes the pink color during the FC portion of the FAS-DPD test, it can over-shoot by up to one drop's equivalent especially if the sample was a faint pink when you added the last drop. So usually, the CC reading in a FAS-DPD test will read from 0 to somewhat less than 0.2 ppm higher for FC and at the same time 0 to somewhat less than 0.2 ppm lower for CC than the "true" reading such that the sum of these two errors equals 0.2 ppm (assuming there is both FC and CC to read -- if the CC is really zero, then an over-shoot won't have any error for CC as it will still read 0). Though one could try using a 50 ml sample size, the resulting resolution of 0.1 ppm isn't really accurate because the pink transition is so faint. How are you getting a 0.1 ppm reading when you titrate?
As for non-chlorine shock (MPS), it reads as TC in the OTO test, FC in the FAS-DPD test and CC in the DPD test.
waterbear may have more info regarding a comparison of titration vs. photometric in his experience.
Why do you care about this 0.2 ppm difference? A CC of 0.2 ppm or less is perfectly acceptable and within existing code standards. Now if you are at a higher CC, then this difference may be more of a problem for you and the titration kit would probably let you pass more readily, assuming the inspector also uses a titration kit.
Richard, Jennie *is* the inspector!
PoolDoc / Ben
Thanks for all the input!
PoolDoc: I would really like to see that news report when you find it.
ChemGeek: The pools I was testing were not previously shocked with MPS. The Palintest photometer we use does have a reagent to offset the effects of MPS on FAC though; it's called OxyStop and we do use it when some of our pools have been recently shocked.
The 0.1 CC reading I got was a very faint pink titration reading. The photometer read 0.3 CC so the difference was about 0.2 CC.
If there is that 0.2 margin of error, it does make sense and I guess that means our photometer is just a bit more accurate. I am concerned about the 0.2 ppm difference because most of my operators still use the titration kit, and I don't want to hold them to eliminating the CC to 0 if they can't even read it on their kits. We advise our operators to eliminate all CC even if it's below legal threshold just so they are aware of it and the CC levels don't get out of hand.