Originally Posted by Amir
I've used Hanna products in the past, and have an extremely low opinion of their quality.
If you have a clear buffer that you can calibrate with, test the buffer with both the Hanna & the phenol red.
One caution: phenol red is NOT stable with chlorine levels above 3 ppm. The phenol red from Taylor, used in the PS200 should be stabilized to 10 - 15 ppm chlorine, but above that will give various false readings.